Help with flow hood please!!
#23818583 - 11/10/16 07:20 AM (7 years, 2 months ago) |
|
|
I have this hepa filter and I need to know what size fan I need I have been up all night trying to figure it out. Finally I give up, so I came here for help...

Thanks for anyone's help that is willing to give me a hand!!
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
|
sandy_vag
Keep Dreaming



Registered: 06/14/16
Posts: 112
Last seen: 5 years, 8 months
|
Re: Help with flow hood please!! [Re: modsolution]
#23818591 - 11/10/16 07:26 AM (7 years, 2 months ago) |
|
|
Who doesn't enjoy reading upside down text...
What size is the filter?
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: sandy_vag]
#23818602 - 11/10/16 07:31 AM (7 years, 2 months ago) |
|
|
Yea sorry about that upside down thing.
size is 24" x 24" x 11.5
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
|
sandy_vag
Keep Dreaming



Registered: 06/14/16
Posts: 112
Last seen: 5 years, 8 months
|
Re: Help with flow hood please!! [Re: modsolution]
#23818959 - 11/10/16 09:39 AM (7 years, 2 months ago) |
|
|
so yo need 400 cfm @ about 1.2" static pressure + prefilter, so arround 1.4" static pressure, minimum.
I think a nice 10 inch inline blower (like hydro ducting blowers) pushing about 800 free cfm would get you there, probably want a speed controller to turn it down a hair.
|
spore-ty



Registered: 01/21/16
Posts: 1,028
Loc: In the bush
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: sandy_vag]
#23818966 - 11/10/16 09:41 AM (7 years, 2 months ago) |
|
|
Quote:
sandy_vag said: so yo need 400 cfm @ about 1.2" static pressure + prefilter, so arround 1.4" static pressure, minimum.
I think a nice 10 inch inline blower (like hydro ducting blowers) pushing about 800 free cfm would get you there, probably want a speed controller to turn it down a hair.
Lol 800 is a bit much but hes right
|
sandy_vag
Keep Dreaming



Registered: 06/14/16
Posts: 112
Last seen: 5 years, 8 months
|
Re: Help with flow hood please!! [Re: spore-ty]
#23818972 - 11/10/16 09:42 AM (7 years, 2 months ago) |
|
|
well the inlines lose their cfm pretty fast at resistance and his filter is higher resistance than most. just a ballpark.
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: sandy_vag]
#23819549 - 11/10/16 12:00 PM (7 years, 2 months ago) |
|
|
Ok im really confused now the below fan marked in green is 985 cfm @ 0.0 SP but only puts out 360 cfm @ 1.25 SP so this fan is not adequate? I do not want to buy the wrong fan or pay more than I should. So thanks for your help all that reply.

I also have the axc200b-es and the axc250-es <---- is 10" but from what I see these are not adequate either. If this is the case, looks like the fan will cost more than its worth. If anyone has any ideas or could suggest a fan, your help would be greatly appreciated.

Thanks to all that replied and will reply!!
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
Edited by modsolution (11/10/16 12:02 PM)
|
sandy_vag
Keep Dreaming



Registered: 06/14/16
Posts: 112
Last seen: 5 years, 8 months
|
Re: Help with flow hood please!! [Re: modsolution]
#23819714 - 11/10/16 12:46 PM (7 years, 2 months ago) |
|
|
Have you tried calling your local hvac heating/ac repair places to see if they have any used 110v squirrel cage fans from customer jobs? I found 2 for cheap or free that way, they wire it up and everything if you ask. Just say you want the biggest 110v squirrel blower you can get for you garage. Should be more than enough power for this filter, most hvac systems have at least a nice 1200+ free cfm fan.
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: sandy_vag]
#23820328 - 11/10/16 04:42 PM (7 years, 2 months ago) |
|
|
No one has an answer? I cant even find the SP for half the fans. How do you get the info that you need to put a flow hood together, if no one gives the SP?
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: modsolution]
#23820440 - 11/10/16 05:34 PM (7 years, 2 months ago) |
|
|
That one is pretgy much like what i have the dayton td1t8 i believe? is what you want check it on amazon
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: Boogieman47]
#23820446 - 11/10/16 05:35 PM (7 years, 2 months ago) |
|
|
That filter at .03 will be about 450-500 somewhere there if you add plenum and prefilter you will be about 1" sp you only need .8 to 1.0 for a proper flowhood
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: Boogieman47]
#23824566 - 11/11/16 11:54 PM (7 years, 2 months ago) |
|
|
OK I think I found the proper blower after a long time searching. This diagram is off the Dayton 1Tdu2. Please check the diagram and let me know guys and gales. I'm counting on all of your experience to let me know if this is what the filter I have requires. I would rather pay a little more and make sure I have what I need. I can get this blower for $130 USD.
Thanks to everyone that has given me a hand!!
Diagram is now below sorry put up wrong pic.
Edited by modsolution (11/12/16 01:37 AM)
|
wtfcrazymofo
foil hater



Registered: 07/26/15
Posts: 1,201
Loc: Colonial alley
Last seen: 14 hours, 36 minutes
|
Re: Help with flow hood please!! [Re: modsolution]
#23824718 - 11/12/16 01:25 AM (7 years, 2 months ago) |
|
|
I got dual 400 cfm in-line fans for my 11.5x2x2. Without the pre filter it is to powerful, but once I popped on the pre filter the lighter test passed and works great.
OP go for 800 or a bit more for yours looks like SP is way high in the upside down pic.
-------------------- If you want to eat->https://www.shroomery.org/forums/showflat.php/Number/8553541 Bag sealers are to bulky (my hood isn't that big) https://www.shroomery.org/forums/showflat.php/Number/28622922
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: wtfcrazymofo]
#23824742 - 11/12/16 01:38 AM (7 years, 2 months ago) |
|
|
Here is the correct diagram...
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: modsolution]
#23824744 - 11/12/16 01:44 AM (7 years, 2 months ago) |
|
|
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: modsolution]
#23824755 - 11/12/16 01:53 AM (7 years, 2 months ago) |
|
|
Thanks wtfcrazymofo for your input. Where I get confused and start to doubt everything is the WG (WG is same as SP right) on the filter says 1.35 max WG. but does not give a min. So I am thinking go higher and I can always cut down the intake air if I have to. That is why I think ill go with the Dayton 1Tdu2.


The airflow at 2000 also confuses me not sure what that means.
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
Edited by modsolution (11/12/16 01:58 AM)
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: modsolution]
#23824778 - 11/12/16 02:08 AM (7 years, 2 months ago) |
|
|
2000 cfm is max think about it like subwoofers orvan amp it says 1500 peak watts thats as high as you will get before it blows i already stated before i have one at 2000 cfm at 1.4 sp which i found the specs and at .3 .4 was about 500 cfm plus you have plenum and prefilter making it around .8 to 1" static pressure water guage water column however youwanna call it so you need 500 cfm at .8" wg but you only need 400 cfm for proper laminar flow........ giving you extra air to bog down and bam problem solved dude
|
modsolution
Stranger


Registered: 07/21/11
Posts: 81
Last seen: 2 years, 10 months
|
Re: Help with flow hood please!! [Re: modsolution]
#23824784 - 11/12/16 02:11 AM (7 years, 2 months ago) |
|
|
noob47 I'm not sure that blower will do it. If I look at the data sheet it only gets about 350 cfm at about 1 WG but from what I understand (which I could be wrong) I need 400 cfm (filter is 2 x 2 x 11.5) at 1.3 WG with a pre filter of .2 which makes me need 1.5 WG total. Please correct me if I am wrong and I will gladly stand corrected.
Thanks again everyone.
-------------------- IF YOU'RE NOT LIVING ON THE EDGE YOU'RE TAKING UP TO MUCH SPACE! I THOUGHT I WAS WRONG ONCE BUT I WAS MISTAKEN!!
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: modsolution]
#23824793 - 11/12/16 02:18 AM (7 years, 2 months ago) |
|
|
No for our needs we need .8 to 1" for proper laminar anything higher and your getting into a whole nother thing ... and that on high at .8 is 500 so at .1 will be 440 or so that gives you more than enough air the plenum gives you .5 sp plus the .2 for prefilter and .3 on your hepa ... the 1.3 is maximum potential for your hood at 2000 cfm so at 400 its like .3 get it?
|
Boogieman47
Let's boogie


Registered: 03/05/16
Posts: 9,712
Loc: Under your bed
|
Re: Help with flow hood please!! [Re: Boogieman47]
#23824797 - 11/12/16 02:19 AM (7 years, 2 months ago) |
|
|
That hepa is nuclear grade
|
|